Voltage measurement on RF cable - RF Cafe Forums
|
Ted1234
|
Post subject: Voltage measurement on RF cable
Posted: Thu Oct 20, 2005 4:03 pm
|
|
|
I have a piece of RG58 going into an into an o-scope
with a 100 MHz BW. The signal is a sinewave (almost)
and I know the power out of my source (10 MHz) is
+10 dBm. Problem is the o-scope is reading about
exactly twice the voltage I expect to see (which
is 2 Vpk-pk). I'm measuring 4 Vpk-pk. Any ideas
what might be going on. Thanks. --- Ted
|
|
|
|
|
kanling |
Post subject:
Posted: Thu Oct 20, 2005 4:08 pm
|
|
|
Colonel |
|
Joined: Mon Jul 18, 2005
4:31 pm Posts: 40 Location: Baltimore, MD
|
I'll bet your scope's input is set to the high impedance
mode. Make sure it is set to be a 50 Ohm load. You
are exactly right to be expecting about 2 Vpp.
|
|
|
|
|
Guest |
Post subject:
Posted: Fri Oct 21, 2005 7:17 am
|
|
|
Agreed, your scope impedance is not 50 ohms. If
you do not have an option for 50 ohms, place a 50
ohm resistor in parallel between clip leads.
|
|
|
|
|
Ted1234 |
Post subject:
Posted: Fri Oct 21, 2005 10:24 am
|
|
|
kanling wrote:
I'll bet your scope's input is set to the high
impedance mode. Make sure it is set to be a
50 Ohm load. You are exactly right to be expecting
about 2 Vpp.
Thanks kanling - you are exactly right
I can't believe I was dumb enough to not
think of that.
Ted
|
|